I have 2 Agilent 1200 DAD systems sat next to each other, bought at the same time, exactly the same 'spec' (exactly the same everything).
One of our analysts came to me and said one of them has 4 times larger peaks for the same runs...
I have run an injection on each system of my own PQ sample using the same column, same phases even the same vial of sample and one system has peaks that are ~ 1/3 bigger than the other!!
My 1st thought was the injector was picking up too much sample but I performed a 'lift' test (weighed a vial of water, carried out 10 x 10ul inj then weighed the vial again) and found that it is picking up the correct amount (10ul).
I spoke to Agilent and they do not know what the cause is...
Any one else had a similar problem?