How can I collect data using HP 6890 Series integrator? I managed to do a plot and when I did a trial injection I observed all the peaks however the 'Plot' function does not provide retention time and area for the peaks which I need for documentation.

When I press 'Start' to begin data acqusition after a manual injection, the integrator stop acquiring data after 1 sec and print 'Stop'. Can someone please tell me why I cannot aquire data for the whole run.

Please someone save me, I served online and read through the manuals still cannot find the solution.

Thanks in advance.

Never heard of a 6890 integrator. Ar you using a stand-alone HP integrator (3396)?

Yes that's the one. It said 6890 on the integrator itself :roll:

Please advise how I can fix the problem.

Thanks.

vliu,

Your run time is set too short. Do a <list><time> to print your timed events table and you should see a Stop command at the 0.05 or something like that. <Del> <Time> <0.05> (or whatever the stop time is) should get rid of that line. Then reset the stop time with <Stop><Time><10.00> to get a 10 minute run.

If this does not work, please re-post and I will dig a little further into the commands (it has been a long time...)

Best regards.

Aicmm,
I checked before,I extended the runtime to 100 mins but still doesn't work. I resetted the integrator ctrl+del+enter just before leaving work. Hopefully will work, will keep u posted.

No luck....same thing, integrator print 'start' then immediately print 'end'. :cry:

vliu,

Please post your list time, list list and list method.

Best regards.

I finally figure out why with the help from Agilent engneerer, it was a cable connection problem. I was told my remote cable is not original and so the connection is not gd. Simple issue and took me so long to resolve...
8 posts Page 1 of 1

Who is online

In total there are 63 users online :: 1 registered, 0 hidden and 62 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Semrush [Bot] and 62 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry