peak separation

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

8 posts Page 1 of 1
I have trying to separate two peak..which i have no idea what kind of molecule are those..casue its unknown from bacterial culture. and i do not have enough to do nmr to know what they are ,..mass spec is not very help ful..i tried c8, c18 and phenyl columns..but they wont separate used acn and water gradient with formic acid..

what should be my next step?,..any advise

Why do you say 'two peaks' if you can't get anything to separate?

If you have access to GCMS I would try shooting it in there. The EI spectral libraries are very good for identigying unknown compounds and if they will fly the separation is generally much better anyway.

I mean two peak because i can separate them very little and they both have different uv max one has 257 and other one has 242 as uv max

Trying different columns is a good way. As you don't know anything about the molecules next tries should be:
- different organic solvent, e.g. methanol
- different pH
Additionally you try to find isocratich conditions. Temperature could also help, especially low temperatures.

Alex

I also would recommend to try another column, but also another chromatographic mode. My first choice would be a HILIC column. HILIC will not work every time, but if one of the 2 compounds is neutral and the other charged, you will get a baseline separation. Hope that will help. Gerhard
Gerhard Kratz, Kratz_Gerhard@web.de

There aren't too many options. Excuse me to repeat:

stationary phase: C18, Phenyl, hydrophilic-embedded, HILIC, IEC, NP, mixed-mode, polymer-based, monolith, hydride silica based, zirconium-based, B1 silica, etc.(sorry, not well categorized)

organic Mobile: ACN, MeOH, THF, acetone (for RP), etc.

Other conditions: temperature, Flow, Ion pair, pH, smaller particle size, longer column, etc.
Excel

at around what % ACN do they come out?
can you give some details of the column you are using, length particle size?

what is the pH of your formic acid solution?
at what concentration? is it formic acid because you use MS/MS?

If your compounds are ionizable (basic or acidic) you best bet is mixed-mode chromatography, where compounds are separated by two mechanisms - reversed-phase and ion-exchange. Here how mixed-mode columns are different from RP or ion-exchange. You are using 2-D approach on a single column:
http://www.sielc.com/Technology_2D_Properties.html

With mixed-mode you can separate structural isomers which are very similar in properties:
http://www.sielc.com/compound_215.html
http://www.sielc.com/compound_308.html
http://www.sielc.com/compound_037.html
8 posts Page 1 of 1

Who is online

In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry