Signal periodically fades
Posted: Thu Aug 12, 2010 2:07 pm
I have a HP 5985 which I have salvaged of eBay thats been stored for about 4 years. I worked on the vacuum and got it to about 1.8 e-6 at which point the galaxy manual says spectra can be done. I repaired a blown RF drive transistor and resistor in the ion focus supply at which point all optics and RF voltages read as they should. The galaxy scan box and main computer I got rid of as it needed a dongle and wrote a brand new GUI for it in visual C++.
At this point I was able to tune a spectrum. I have large signals at 28 and 32 (1/2 of full log-scale, correspondig to my vacuum I suppose. Trouble is after I let in a sample (I use acetophenone which has about the same vapor pressure as PFTBA) the sample peaks appear, and then the whole thing fades, almost all peaks disappaer - they do not shift - just colapse into the noise where they were. The source current stays as programmed - as otherwise the MS reports a fault. After I close off sample, the air peaks reappear taking about 1 min, and oscillating in and out before establisng themselves - but always at the same values 28 - 32. I guess this excludes the RFPA, and source cleaning is in order? Has anyone observed such behaviour.
Another question I have is to do with the EM. I read after two years it should be replaced. But what are the indications for this? My baseline (at no source current) is very low and flat, while the background N2 and O2 is enough to give 1/2 full log-scale - so is there any reason why one would want to replace the EM? I did need to tune a low x-ray voltage to get a spectrum - about 50V. At about -200V xray voltage the spectrum disappears completely and is replaced by off-axis noise (its not thermal noise as the noise disappears completely when ionisation current is turned off). I do not know if this is normal.
A third question is to do wth the pumps. I turn them off after work to extend their life and the life of the motor. Is that generally done? The motors are old - the E2M2 looks like its from the 1970's.
At this point I was able to tune a spectrum. I have large signals at 28 and 32 (1/2 of full log-scale, correspondig to my vacuum I suppose. Trouble is after I let in a sample (I use acetophenone which has about the same vapor pressure as PFTBA) the sample peaks appear, and then the whole thing fades, almost all peaks disappaer - they do not shift - just colapse into the noise where they were. The source current stays as programmed - as otherwise the MS reports a fault. After I close off sample, the air peaks reappear taking about 1 min, and oscillating in and out before establisng themselves - but always at the same values 28 - 32. I guess this excludes the RFPA, and source cleaning is in order? Has anyone observed such behaviour.
Another question I have is to do with the EM. I read after two years it should be replaced. But what are the indications for this? My baseline (at no source current) is very low and flat, while the background N2 and O2 is enough to give 1/2 full log-scale - so is there any reason why one would want to replace the EM? I did need to tune a low x-ray voltage to get a spectrum - about 50V. At about -200V xray voltage the spectrum disappears completely and is replaced by off-axis noise (its not thermal noise as the noise disappears completely when ionisation current is turned off). I do not know if this is normal.
A third question is to do wth the pumps. I turn them off after work to extend their life and the life of the motor. Is that generally done? The motors are old - the E2M2 looks like its from the 1970's.