7673B/5972/5890II. with chemstation A1701G BA 1.0.0. Ver.
The problem is: After the injector injects the sample the GC would not respond (ramp the temperature) and both the red and the green light in the GC go off and the tray arm wont take the vial back to place it in the tray. Also there is no respond for the chemstation(perinjection state).

This may be very obvious, but are there any lights on the injector tower and injector controller box showing. The tower error light will be a series of flashing lights. Count the number of "flashes" in the series.

You might also check to see what you have selected in the software for the start trigger.

faris,

Is the remote cable connected? Should be a cable running from the GC (upper inside right cover - remote position) to the back of the autosampler controller and probably to the MS as well (sorry, been a while.)

Best regards.

I have seen that before, it is usually caused by the tower.
4 posts Page 1 of 1

Who is online

In total there are 70 users online :: 1 registered, 0 hidden and 69 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 69 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry