Ammonia by HPLC?
Posted: Mon Aug 09, 2010 2:38 pm
by HPLC_Help
Is HPLC a good way to measure ammonia? What kind of detector would work? RI?
What about GC? Would that be better?
Or is there a better method altogether.
I'm monitoring ammonium usage during a biological reaction.
Thank you for pointing me in the right direction!
H_H
Posted: Mon Aug 09, 2010 3:14 pm
by oscarBAL
Hello H_H; ammonia can be analyzed either by HPLC (using ion exchange) or GC; but sometimes and depending of your concentration; steam destillation and Titration could be easier solution; if your problem is sencibility you can also use Selective Ammonia electrodes or colorimetric method; it is necesary to have in mind that these tree methods are sencitive to volatil amines also but work easier and faster.
Posted: Tue Aug 10, 2010 6:26 am
by Markus Laeubli, Metrohm
Hallo HPLC_Help
oscarbal is absolutely right. Ion Chromatography, titration or ISE all work in their respective concentration ranges and sample matrix conditions. For hgher concentrations I would go for Titration / ISE. For lower concentration and/or amines around, IC will be the preffered method.
Posted: Wed Aug 11, 2010 12:08 pm
by Phycal
Also can use the Nessler reagent and do it spectrophotometrically.
Posted: Sun Aug 15, 2010 5:08 pm
by praveenpaliwal
Try Inertcap column for volatile amine u can get good responce not only for ammonia but other volatile amine also on GC-TCD
Praveen
Posted: Mon Aug 16, 2010 9:36 pm
by SIELC_Tech
here is method for analysis of ammonia using Primesep mixed-mode column and ELSD detector. You can analyze non-UV active amines using this approach:
http://www.sielc.com/compound_005.html
Contact me if you have questions