After changing the column, inlet sleeve, seal, and septum I am still having the same problem. A while back I was talking with a guy from Agilent and he said that the newer MS's were more sensitive (of course), but that this was possibly causing some type of ion competition in the detector itself. So, what that basically told me was that as instrument technology makes the sensitivity better...the chromatography gets worse...

This was not very uplifting news. Does this seem at all possible?
Recap of problem...after new column: I'm using three internal standards and the last one is coming out the worst. 1,4-dichlorobenzene-d4. In my 10ppb standard the response is 153000 and in my 200ppb standard it is 247000. Then when my samples start it goes down to 98000 and remains pretty consistent.
Peter,
In addition to the moisture and hydrocarbon traps, I found a "Superior Helium Purifier" in line also. It was strapped to the back and I overlooked it. I had our gas lab check the He for impurities and it checked out fine, however, they can only see to 5ppm.
Bigbear,
I did not solvent wash the inlet...just changed everything out. This has been a problem since the system was set up...brand new, so I'm trying to think outside of active sites. Yes, the P&T is hooked up to the injector. I'll have to do a direct inject of MeOH sometime next week. I still have to pump out data even with mediocre results.

Oh, and most of my aromatics are the problem...
Thank you guys!