Max. ion pair conc.
Posted: Mon Jan 24, 2005 1:51 pm
Is there a max. working limit to the amount of ion pair conc. in the mobile phase that you can readily work with.
Our present method uses a 20mM SDS + 0.05% TFA as MP A with a 5 cm column. However, we'd like to increase our LOD for the basic solute... presumably by simply injecting a larger volume. By gaining more retention on our column by adding more ion pair reagent will the system be able to tolerate a larger injection volume or should we simply increase the length of our column to increase column injection volume. (Please note that our sample diluent is fixed and cannot be changed thus eliminating on-column conc. techniques).
Finally for best results with ion pair reagents, is it best to pre-mix your mobile phase with the organic portion or is it just as well to let the pump mix online? I have tried using a C16 sulphonate but it didn't dissolve in the aqeous portion of the mobile phase. Perhaps I should have premixed with the organic component?