CP PoraBond Q - Poor Tailing Factor
Posted: Sun Aug 01, 2010 6:22 pm
Dear colleagues,
I have run ethanol determinations from an aqueous solution containing polysorbate 80 for several years without problems. During the past year, the Tailing Factors for ethanol and propan-1-ol (internal standard) are highly variable throughout the run and often >>1.5 (in-house limit), sometimes >2.0 Also, peak areas can be highly erratic.
Conditions:
Column 25m x 0.32mm x 5um
Col T = 150C
Inj T = 150C
Det T = 240C
Carrier gas Helium
Flow 1ml/min
Split 1:50
Injection volume 1uL or 0.5uL (depending on application)
I have recently adopted a policy of regular cleaning of the split liner and am using a liner with a 'cup' near the base without silanised glass wool.
I have also cut about 20cm from the inlet end of the column.
The chromatography has improved slightly but is still not as good as it used to be.
Any suggestions for further improvement?
Conditions:
Col T = 150C
Inj T = 150C
Det T = 240C
Carrier gas Helium
Flow 1ml/min
Split 1:50