In the valiadation of a stability-indicating assay and impurities methods,
what acceptance criteria do you all normally apply to robustness results? Also, do you normally evaluate wavelength as part of robustness testing, and if so, what range do you usually evaluate (e.g., +/- 2 nm)?

Thanks!
Hi all,

In the valiadation of a stability-indicating assay and impurities methods,
what acceptance criteria do you all normally apply to robustness results? Also, do you normally evaluate wavelength as part of robustness testing, and if so, what range do you usually evaluate (e.g., +/- 2 nm)?

Thanks!
WV is not included. It should be calibrated routinely in PM, PQ.
I was wrong, how about 2% variation?
2 posts Page 1 of 1

Who is online

In total there are 63 users online :: 2 registered, 0 hidden and 61 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Majestic-12 [Bot], Semrush [Bot] and 61 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry