Ghost Peak in HPLC EC
Posted: Fri Jul 30, 2010 4:18 pm
I have been trying to run HPLC with electrochemical detection for dopamine and serotonin in microdialysates but have been getting a ghost peak that interferes with dopamine. I have been unable to get rid of this peak and still can't figure out where it's coming from. Details about the protocol and steps I've taken to get rid of the peak are listed below. Any help in getting rid of the ghost peak will be much appreciated!
System:
Shimadzu LC 20 AD
Rheodyne manual injector (have tried steel 7125 and PEEK 9125)
Phenomenex Kinetex C18 column (100x2.1)
BAS radial flow cell for electrochemical detection
Also have guard column and pulse damper
Protocol:
Mobile phase is 150 mM NaH2PO4, 4.8 mM Citric acid, 3mM SDS, 50 uM EDTA, 11 % MeOH, 17% ACN, pH 5.6 with NaOH
Electrode set at 550 mV
Flow rate is 0.2 mL/min
Information about Ghost Peak:
Shows up regardless of what I inject (mobile phase straight from reservoir, aCSF, HPLC grade water, etc.)
Does not appear when flipping injector to load and back to inject
Have tried 4 injectors and sample loops, rotor seals made of vespel or tefzel, many different hamilton syringes, several columns, and several guard columns but peak is still there
Peak size changes with injection volume (so larger at 20 uL injection vs. 5 uL)
Peak size changes with electrode potential suggesting it is an electroactive compound
System:
Shimadzu LC 20 AD
Rheodyne manual injector (have tried steel 7125 and PEEK 9125)
Phenomenex Kinetex C18 column (100x2.1)
BAS radial flow cell for electrochemical detection
Also have guard column and pulse damper
Protocol:
Mobile phase is 150 mM NaH2PO4, 4.8 mM Citric acid, 3mM SDS, 50 uM EDTA, 11 % MeOH, 17% ACN, pH 5.6 with NaOH
Electrode set at 550 mV
Flow rate is 0.2 mL/min
Information about Ghost Peak:
Shows up regardless of what I inject (mobile phase straight from reservoir, aCSF, HPLC grade water, etc.)
Does not appear when flipping injector to load and back to inject
Have tried 4 injectors and sample loops, rotor seals made of vespel or tefzel, many different hamilton syringes, several columns, and several guard columns but peak is still there
Peak size changes with injection volume (so larger at 20 uL injection vs. 5 uL)
Peak size changes with electrode potential suggesting it is an electroactive compound