Recently we need to do performance check for the GC-MS (6890-5973N, EI mode). But no one in our lab have experience about using Chemstation to perform s/n check.
I tried to inject 1ul of 1ppb OFN solution, scan from 50-300amu, click Instrument Control --> Qualify --> Sensitivity Check.
After around 6 min, it come out a report but no peak of m/z 272 at all. And the s/n is 15.
May I know:
1. The normal range of s/n of 5973N? (some data sheet said 60:1, some said 20:1, how about yours GC-MS?)
2. Is my step correct? (Instrument Control --> Qualify --> Sensitivity Check)
3. I found that in the Data Analysis --> Tools --> Sensitivity. What is the difference between this and pt.2?
4. If no m/z 272 found. What action should I do? (e.g. clean ion source?)
Thanks for your help