HP7694 carousel adjustment

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

5 posts Page 1 of 1
Our 7694 Headspace sampler carousel is out of adjustment. According to the manual I have, it's says to "call Agilent service"...

In the past I have adjusted the optical sensor position under the elevator area, but I did that Monday and it's back in my shop out of alignment again after only 10 to 15 vials through it. (I don't have service manuals/schematics, nothing appears to be loose, the belt is tight, does not appear to be excessively worn, but maybe it should be replaced...???) Right now I'm at the end of the adjustment for the optical sensor.

The 7694 is needed "yesterday" so am hoping someone here knows the proper alignment procedure or other suggestions.

Thanks!!

dersh.
Maybe this will help? Pages 97 to 101 are adjusting steps.

"If the elevator rod and bushing do not line up exactly under the hole in the
carousel cylinder, call Agilent service.
"

That's the same manual I have,

dersh.

I don't have a 7694, so I can't give specific advice, but I can try to make comparisons to the instrument I do have, the Tekmar HT3. If I am adjusting the carousel, elevator, mixer, etc. there is usually an optical sensor and a "flag" that passes through the sensor. Perhaps you can move the flag instead of the sensor itself to get some extra adjustment?

Have you tried calling Agilent? They are very helpful...I've made three calls this week for a 5975 that has a terrible mass-shift error.

Good luck.
Thanks Schmitty, I've been out of town for a bit...

I had been aligning the optical sensor (not the wheel), but I'm now at the end of adjustment. So far it has been staying aligned, but when/if it needs further adjustment, I have no where to go.

I did call Agilent 2 weeks ago, the person I talked to said they would need to find someone more knowledgeable to help, I still haven't heard from them.

I'm guessing I'll have to pull apart the carousel assembly and move the optical wheel then realign the optical sensor - this appears to be a major undertaking (I have no service manual), unless someone has a better "fix".

Thanks again, appreciate the help,

dersh.
5 posts Page 1 of 1

Who is online

In total there are 12 users online :: 1 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Manmohan and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry