Pepter » Fri Jul 23, 2010 10:52 pm
Once again.
1) You can see parent ions on scan mode (MS1 Scan ?) when the analytes go through column
2) Cant see any of transition ions from MRM mode when the standards go through column
3) MRM mode and all transitions works fine with manual injection
Do you use the same method for both manual and autosampler injection ?
Is the amount of analytes the same for both injection types ?
All transitions that You list are in one time segment ?
You write that you double-checked MRM by re-run standards with ion transitions... so it was manula injection ?

give us some print screens of your acquisition pan
If you solve the problem also write please

sounds mysterious .