				epc oscillation
				Posted: Mon Jul 26, 2010 1:08 pm
				by chromtech
				I have a 5890 series 2 that has been upgraded with an epc system.  With the purge valve off it controls column head pressure fine.  When the valve is opened for split injection the pressure oscillates above and below the set point.  Is this a symtom of an eprom mismatch?
col. head pres: 12.6
split flow: 120ml/min
			 
			
					
				
				Posted: Mon Jul 26, 2010 3:19 pm
				by Peter Apps
				Check that the pressure feeding the GC is high enough, and that there are no restrictions in the lines.
Peter
			 
			
					
				epc oscillation
				Posted: Mon Jul 26, 2010 6:14 pm
				by chromtech
				carrier pres. is @85 psi to the gc regulator,  no kinks or obstructions.
			 
			
					
				
				Posted: Mon Jul 26, 2010 9:22 pm
				by bhuvfe
				what is the spread of the oscillation?
Maybe check for leaks at the inlet (starting from the septum).
			 
			
					
				
				Posted: Tue Jul 27, 2010 6:37 am
				by Peter Apps
				If you have a trap between the inlet and the back pressure regulator (I am assuming that you have the usual kind of plumbing for a split splitless inlet) it may be blocked or significantly restricted.
Peter
			 
			
					
				5890 epc oscillation
				Posted: Wed Jul 28, 2010 12:46 pm
				by chromtech
				I have been all over the gc with a sniffer, can't find any leaks.  Also the oscillation doesn't seem to start until the instrument is connected via HPIB to chemstation software.  There is no trap, don't ask me why.
Thanks for any help in this forum, it is greatly appreciated
			 
			
					
				
				Posted: Wed Jul 28, 2010 2:58 pm
				by Peter Apps
				Check that all the settings in Chemstation match reality;  column dimensions, carrier gas, pressure units, septum purge flow and anything else that I have forgotten.  Also it can't do any harm to measure the actual flows with an independent flow meter.
Peter
			 
			
					
				epc oscillation
				Posted: Wed Jul 28, 2010 5:10 pm
				by chromtech
				after substition found that the septum purge regulator was faulty.
thanks again
			 
			
					
				
				Posted: Thu Jul 29, 2010 6:13 am
				by Peter Apps
				Thanks for the feedback
Peter