Hi, can ethanol be analysed by HPLC with UV detection at low wavelength

Thanks

Martina

No. It is not possible. Try GC.

I doubt you could every get that to happen without massive amounts of work.

You would be better off looking to capillary GC with an FID.

You can analyze Ethanol by HPLC but using IRD.

Yeah... you need the RID. If using the Biorad aminex column then
ethanol comes out around 22 minutes at 1 mL/min.

Doubt not. It can be done pretty simply with an RI detector & I've done it, though it was a long time ago. Get an RI, check out Shodex' website, and have fun. An old Waters 410 RID will be cheap and they run forever. Patience will be required to obtain a stable baseline, but once you have one, you'll be fine.
6 posts Page 1 of 1

Who is online

In total there are 15 users online :: 1 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry