How can we reduse noise baseline in the parameters?
We are experiencing noise like the picture above:
Advertisement
Discussions about GC and other "gas phase" separation techniques.
Absolutely. Many things can cause stuff like this: leaky septum, plugged syringe needle, detector gasses, too little analyte, leak at inlet.However, you should not be using the ChemStation to solve an instrument problem, such as a dirty detector or leak in your system, which may be causing your noise.