shaun78 » Tue Jul 20, 2010 12:31 pm
It depends what you are after. If the solvent savings is important to you, then scale it down. If you really do not care, then leave it the same.
If you do scale down the column, you should maintain similar chromatograpy as long as you closely follow the method transfer calculations. If you are unfamiliar with those, see the following article:
http://www.restek.com/restek/images/ext ... AN1033.pdf
You are going to have to change the flow rate and the injection volume.
Back when the acetonitrile scare was in high gear I scaled down a good number of our analytical methods from a 4.6 mm column to a 3.0 mm column to cut solvent consumption. I rarely had troubles when I did so as long as I adjusted column flow to maintain linear velocity and injection volume to maintain column load.