FID Problem in Varian 450
Discussions about GC and other "gas phase" separation techniques.
6 posts
Page 1 of 1
i am using Varian 450-GC with capillary Column, the problem is that there is flame out problem in FID but once the sample is injected it was ok but as soon as sample completes it gives warning of fleme out
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Before you go any further you need to check whether the flame really is out, or whether it is burning but the electronics are giving a false error message. To check whether the flame is burning hold a shiny, cold metal or glass surface above the detector and look for condensation. If you see condensation the flame is burning.
If the flame is not burning then measure all the flow rates (with a separate flow meter not the readouts on the GC), including the carrier gas.
Check that you do not have a programme for the FID that might make it shut down at the end of a run.
How do you get it to be ready at the start of a run ?
Peter
If the flame is not burning then measure all the flow rates (with a separate flow meter not the readouts on the GC), including the carrier gas.
Check that you do not have a programme for the FID that might make it shut down at the end of a run.
How do you get it to be ready at the start of a run ?
Peter
Peter Apps
-
- Posts: 24
- Joined: Mon Jun 21, 2010 5:38 pm
after several attempts the Flame is on & ready for some seconds & we inject sample but it will also flame out during run but fortunately there r 6 minutes before first peak & it is not flame out till 35 minutes run
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
And the results of the tests and flow rate measurements are ...... ???????
Peter Apps
-
- Posts: 24
- Joined: Mon Jun 21, 2010 5:38 pm
there is no leakage i have changed the Burner tip but after that fid works well but my sample last peak is very large what is the reason while other peaks r normal plz help
-
- CE Instruments
Check the auto zero, Flame out is usually indicated when the background level when lit is below a certain threshold. This can occur if the GC will accept a negative value as zero when not lit. Are you flow programming ? increase of flow , decreases the baseline getting you below flame out threshold ?
6 posts
Page 1 of 1
Who is online
In total there are 13 users online :: 1 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 12 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 12 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science