In red is the good baseline, in blue is a later injection with the pump doing it's erratic 'clicking' thing. This is an injection of the same vial about 15 hours apart.
I got the pump to stop making the extra noise by opening the purge valve and purging each channel for 10 minutes at 5 mL/min, then resetting to 1 mL/min and resuming the run and evaluating the baseline. The extra 'clicks' went away, and the baseline returned to normal.
Has anyone heard their 1100-series pump making a noise like this? I didn't notice any sort of wild pressure fluctuations - in fact, a pressure trace overlay is almost identical - flat and regular - even though the baseline on one injection is good and the other is noisy. I'm concerned I'm seeing something that is indicative of an impending catastrophic failure, and I'd like to head it off at the pass (even though I hate that cliche) before it becomes an unserviceable issue.