Halfing the flow rate results in doubling the peak area – not because you elute more but because you elute the same amount over longer time. If you compare the peak heights at flow rate 1 mL/min and 0.5 mL/min you’ll observe approximately the same heights but the peak width with the 0.5 mL/min flow rate will be the double of that when the flow rate is 1 mL/min.
That leads me to the sensitivity thing: Acquiring larger peak area does not mean improving the sensitivity. The sensitivity is peak height dependent (actually dependent on the signal to noise ratio, but in your case you need to think peak height).
Finally, if you inject particles or whatever, that causes pressure increase, it’s not only a problem for narrow bore coluns but for wider diameter columns as well – it just take longer time for the bill to arrive to you.
Best Regards
Thanks Danko for your reply, for the sensitivity issue I might have another opinion, first I see the doubling in both peak area and peak height, again the retention time is almost constant because the eluting factor here is %B (remember the gradient is constant),, have you tried making this direct comparison before and not notice this? I think your explanation may apply mainly to isocratic methods?
as for our sample impurities, of course we don't inject particles

, it's just samples heavily loaded with dissolved electrolytes,vitamins, peptides and peptones mixtures in addition to our analyte of interest, so I just assume a bigger column would have a higher capacity..
has anyone else tried this flow rate lessening and observed increased peak area and height, using the same gradient?