Setting dwell time in GCMS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
I'm setting up a GCMS method on Agilent instrument. I wonder if the dwell time needs to be set the same for all ions within a group and between different groups. The Agilent Chemstation manual recommends same dwell time be set but did not specify within a group or between a group. Is it a general practice to optimize dwell time for all ions by experiments?
Please kindly advise. Thanks.

Like my laywer says: It all depends.

If you are looking at ions with about the same abundances, the same dwell time for for each, across the board should be just fine. If you need a bit more sensitivity for a partiulcar ion, you can dwell longer on that ion. Just remember: what you give to one, you take away from the others.

Thank you very much. Do you mean I can set the dwell time as high as 100 msec for some ions as long as the number of cycles / peak comes out to be within 12-25 per peak as suggested by the Agilent manual? I guess I have to test several settings on the samples.
3 posts Page 1 of 1

Who is online

In total there are 27 users online :: 1 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider] and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry