The ICS-3000 SP's alarm was red.
Trouble shooting guide: Failed pressure transducer top or bottom pump.
How do I fix it? thanks.
Jady
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Do you have an isocratic or gradient pump? If it's isocratic, then your pump has a 2-way eluent flow valve that will not open unless you turn the pump flow on via normal operation or priming. Hence, you would not be able to pull much liquid into the syringe with this valve closed. When you tried pulling liquid out with a syringe, was the prime or flow light on at the top of the pump's front panel lights.Thanks Tom and John.
John: the way you mentioned I do every time before I run HPLC, everyday.
No liquid came out from the clean tube when I did prime pump. It seams the fluid in the clean tube goes back and forward. I opened the priming valve, and tried to pull liquid out, failed.