Chelators in LC-MS
Posted: Wed Jul 07, 2010 7:55 pm
Hello,
I am trying to convert an LC-UV method to an LC-MS method in order to improve sensitivity. The problem is the original analyte is a chelator, and so the LC-UV method used EDTA in the mobile phase. Here are the mobile phase parameters:
65:21:14 0.01M phosphate buffer (pH 3.0):Methanol:ACN with 2mM EDTA
Discovery HS F5, 250mm*4mm, 5um, 40 deg C
Flow rate 1mL/min
This procedure is cited in Journal of Pharmaceutical and Biomedical Analysis 43 (2007) 1343-1351, Development and validation of HPLC-DAD methods for the analysis of two novel iron chelators with potent anti-cancer activity, by Mrkvickova et al.
The compound I am analyzing is also a chelator.
I'm aware of the issue with phosphate buffer in MS so I am going to select a different buffer. However, my main issue is trying to replace the EDTA. I've used heptafluorobtyric acid as an MS-friendly ion pair reagent in the past, can it be used effectively as a competitor for chelation (like EDTA) so that my analyte doesn't pick up an ion?
Thanks,
Chris
I am trying to convert an LC-UV method to an LC-MS method in order to improve sensitivity. The problem is the original analyte is a chelator, and so the LC-UV method used EDTA in the mobile phase. Here are the mobile phase parameters:
65:21:14 0.01M phosphate buffer (pH 3.0):Methanol:ACN with 2mM EDTA
Discovery HS F5, 250mm*4mm, 5um, 40 deg C
Flow rate 1mL/min
This procedure is cited in Journal of Pharmaceutical and Biomedical Analysis 43 (2007) 1343-1351, Development and validation of HPLC-DAD methods for the analysis of two novel iron chelators with potent anti-cancer activity, by Mrkvickova et al.
The compound I am analyzing is also a chelator.
I'm aware of the issue with phosphate buffer in MS so I am going to select a different buffer. However, my main issue is trying to replace the EDTA. I've used heptafluorobtyric acid as an MS-friendly ion pair reagent in the past, can it be used effectively as a competitor for chelation (like EDTA) so that my analyte doesn't pick up an ion?
Thanks,
Chris