Hi I hope someone can help me. I have to calculate repeatability for finished product (total impurities).
I ran 6 samples (spiked at the impurity limit) and three controls (unspiked)

Should the total impurities recovered for the run be the sum of total impurities for the six samples - the average total impurities in the control samples or is it the sum of total impurities in the six samples+the average total impurities in the control samples.

Thanks

Martina

Do whatever your protocol or SOP says!

If you have no protocol or SOP for this, then, for heaven's sake, write one to avoid the next person having the same problem/question.

Repeatability, by definition, is run on the same sample. Therefore, for this purpose, you ignore your "controls", since they have different levels from the spiked samples.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 18 users online :: 2 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: John Guajardo, Semrush [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry