comparison between CE and UPLC
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
10 posts
Page 1 of 1
I am thinking using one very powerful technique for comprehensive analysis. right now, candidates has been narrowed down to uplc and CE. Could you help me out to choose one of them. are their plate count similar to each other? analysis time? any comprehensive paper on comparison between these two techniques? thanks a lot.
-
- Posts: 366
- Joined: Wed Feb 02, 2005 9:34 pm
Well, right off the bat you can eliminate CE as being in any way comprehensive; because it only works for charged analytes.
-
- Posts: 835
- Joined: Thu Apr 14, 2005 7:00 am
when you say comprehensive, then for what? for what application?
CE is very popular and a strong solution for the bio world,
what type of detection do you intend to use as well will define how comprehensive you are capable of getting for your field of interest
both UHPLC and CE instruments are very sensitive ones, so it is good in one sense and bad in the other one,
check technical support and learn maintenance needs,
if you are told by the vendors not to worry then stop and look for end users to speak to.
No one instrument in a given configuration can give you the solution for everything in the universe.
the better the solution the more expensive it will be,
define your needs carefully
CE is very popular and a strong solution for the bio world,
what type of detection do you intend to use as well will define how comprehensive you are capable of getting for your field of interest
both UHPLC and CE instruments are very sensitive ones, so it is good in one sense and bad in the other one,
check technical support and learn maintenance needs,
if you are told by the vendors not to worry then stop and look for end users to speak to.
No one instrument in a given configuration can give you the solution for everything in the universe.
the better the solution the more expensive it will be,
define your needs carefully
-
- Posts: 1889
- Joined: Fri Aug 08, 2008 11:54 am
following on from adam's point, UPLC using conventional RP media is awful for exactly the things at which CE excels. Sugar phosphates, nucleotide phosphates etc. are a real struggle even with hilic, and hilic isn't the best choice if you have non-polar things present. Comprehensive is tricky by any technology (even assuming you can extract everything comprehensively).
Adam is also wrong: CE can be applied to uncharged analytes. If appropriate detergents are included, totally uncharged molecules can bind to detergent micelles and be dragged along because the detergent is charged. Uncharged things then separate on the basis of partitioning between solvent (which moves with the electroosmotic flow if any exists) and micelles that are being dragged elecrophoretically. It seems a strange thing to do, but the advantage is that there is still no pressure-driven bulk flow, so there is no laminar-flow distortion, and the peakwidth remains very narrow.
Adam is also wrong: CE can be applied to uncharged analytes. If appropriate detergents are included, totally uncharged molecules can bind to detergent micelles and be dragged along because the detergent is charged. Uncharged things then separate on the basis of partitioning between solvent (which moves with the electroosmotic flow if any exists) and micelles that are being dragged elecrophoretically. It seems a strange thing to do, but the advantage is that there is still no pressure-driven bulk flow, so there is no laminar-flow distortion, and the peakwidth remains very narrow.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Jiang925, I believe that you work primarily with proteins, right? In very general (and oversimplified!) terms, the higher the molecular weight, the better the efficiency (plate number) of CE and the worse the efficiency of LC. That's one side of it. The other side is that CE has sensitivity issues (with UV detection) and injection reproducibility issues.
If you are looking for a technique to check your LC prep purification, CE has the advantage that it is pretty much orthogonal to LC in terms of selectivity.
If you are looking for a technique to check your LC prep purification, CE has the advantage that it is pretty much orthogonal to LC in terms of selectivity.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 214
- Joined: Wed Dec 30, 2009 4:51 am
thanks a lot. guys.
I am working on small molecules from food rather then on protein. It is very similar to metabolic profiling, which try to find out which compound is responsible for different flavor between two products.
I am working on small molecules from food rather then on protein. It is very similar to metabolic profiling, which try to find out which compound is responsible for different flavor between two products.
Jiang925, I believe that you work primarily with proteins, right? In very general (and oversimplified!) terms, the higher the molecular weight, the better the efficiency (plate number) of CE and the worse the efficiency of LC. That's one side of it. The other side is that CE has sensitivity issues (with UV detection) and injection reproducibility issues.
If you are looking for a technique to check your LC prep purification, CE has the advantage that it is pretty much orthogonal to LC in terms of selectivity.
-
- Posts: 214
- Joined: Wed Dec 30, 2009 4:51 am
I am afraid that MECK technique cannot couple to MS, am I right?
following on from adam's point, UPLC using conventional RP media is awful for exactly the things at which CE excels. Sugar phosphates, nucleotide phosphates etc. are a real struggle even with hilic, and hilic isn't the best choice if you have non-polar things present. Comprehensive is tricky by any technology (even assuming you can extract everything comprehensively).
Adam is also wrong: CE can be applied to uncharged analytes. If appropriate detergents are included, totally uncharged molecules can bind to detergent micelles and be dragged along because the detergent is charged. Uncharged things then separate on the basis of partitioning between solvent (which moves with the electroosmotic flow if any exists) and micelles that are being dragged elecrophoretically. It seems a strange thing to do, but the advantage is that there is still no pressure-driven bulk flow, so there is no laminar-flow distortion, and the peakwidth remains very narrow.
-
- Posts: 214
- Joined: Wed Dec 30, 2009 4:51 am
but for small metabolites, are their efficiency quite similar. and I think CE also does not have enough peak capacity even they use MECK technique.
Jiang925, I believe that you work primarily with proteins, right? In very general (and oversimplified!) terms, the higher the molecular weight, the better the efficiency (plate number) of CE and the worse the efficiency of LC. That's one side of it. The other side is that CE has sensitivity issues (with UV detection) and injection reproducibility issues.
If you are looking for a technique to check your LC prep purification, CE has the advantage that it is pretty much orthogonal to LC in terms of selectivity.
-
- Posts: 1889
- Joined: Fri Aug 08, 2008 11:54 am
I certainly wouldn't put detergents into an MS... ion pair reagents give me the heebie jeebies already!
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
If the flavours are odours use GC, with a sniff port, and then heart cut the part of the chromatogram where the target odour elutes.thanks a lot. guys.
I am working on small molecules from food rather then on protein. It is very similar to metabolic profiling, which try to find out which compound is responsible for different flavor between two products.
If the flavours are tastes you are stuck with liquid phase separations.
Peter
Peter Apps
10 posts
Page 1 of 1
Who is online
In total there are 30 users online :: 0 registered, 0 hidden and 30 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 30 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 30 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science