I have to run a supplied method with 6mM NaCl in m.p. on Agilent 1200RRLC but the manual says to avoid NaCl because it is corrosive. What is the NaCl for - it is not a bio-molecule analysis? Has anyone any experience of running NaCl on a non-inert pump (i.e. stainless steel flowpath)? How high is it safe to go with the concentration?
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
I've run 0.3 M NaCl in one of the eluents (the strong one in IEx) for several years with no significant problems to observe. As long as pH isn't acidic it's save.
What does it do in you're eluent? I'm not sure. There might be a rationale for that, but it might be also a misconception.
NaCl is a very potent eluting salt in IEx chromatography. Also in low concentrations it's a viable chaotropic agent.
Best Regards
P.S. Whatever the reason, it's a good routine to wash the system for salt containing mobile phase, following use of that type of eluents. Especially if the system is not to be used for several days, weeks etc.
What does it do in you're eluent? I'm not sure. There might be a rationale for that, but it might be also a misconception.
NaCl is a very potent eluting salt in IEx chromatography. Also in low concentrations it's a viable chaotropic agent.
Best Regards
P.S. Whatever the reason, it's a good routine to wash the system for salt containing mobile phase, following use of that type of eluents. Especially if the system is not to be used for several days, weeks etc.
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 33
- Joined: Fri Jun 25, 2010 7:02 pm
NaCl is corrosive because 316SS can become pitted by it. That said, washing the system everyday to remove the salt is important for longevity of the system.
-
- Posts: 10
- Joined: Mon Jun 08, 2009 1:02 pm
Thanks for comments.
4 posts
Page 1 of 1
Who is online
In total there are 36 users online :: 1 registered, 0 hidden and 35 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 35 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 35 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science