How to increase sensitivity for Waters 2996 PDA

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Dear All,

How to increase sensitivity for Waters 2996 PDA, I'm not sure that I can or not.

Thanks.

Depends on what you mean by "sensitivity".

If you are referring to the output setting (how many absorbance units full scale), that is usually done in the data system rather than the detector.

If you are referring to limit of detection or limit of quantitation, then it's a question of signal/noise ratio. You need either more signal (inject more sample and/or make sure you are measuring at an absorbance maximum of your analyte) or less noise (a new lamp, increasing the bandpass, setting a longer time constant).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Thank you for your kind support. :lol:
3 posts Page 1 of 1

Who is online

In total there are 15 users online :: 1 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry