Can I use normal HPLC for SFC?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi all,

Are there any special features of SFC hardware or we can use normal HPLC instruments for SFC?

Thanks.

I'm not familiar with the system mentioned in this press release (Aurora SFC Fusion A5), but Agilent claims...

"The Aurora SFC Fusion A5 connects to a standard HPLC system and transforms it into a complete SFC system without requiring additional hardware or software, disabling features or compromising performance. Installation is completely reversible, so the LC can be used in either SFC or HPLC operating mode. The Aurora SFC Fusion A5 requires only one instrument for normal and reversed phase separations and works with standard (not SFC) grade CO2 for low operating costs."

http://www.agilent.com/about/newsroom/p ... 09016.html

Generally speaking, unmodified HPLC systems are not suitible for SFC. It is not as simple as putting together a list of "change this to that" either but it is possible (I'v done it), I will try to hit some of the highlights.

1. Pump heads need to be cooled to remove the heat generated by the compression of the carbon dioxide.

2. Some materials in some systems are not compatable with carbon dioxide (teflon comes to mind). Some seal materials tend to shrink causing leaks. Some check valve materials are prone to this as well.

3. Some system use manifold valves (like GPV's in Quartanary systems or A1/A2 switching in high pressure binary systems). These valves are on the low pressure side of the system and usually cant take more than 100 psi with out leaking. The input pressure of CO2 is ~ 800 psi.

4. Some pumps use pre-compression algorythyms in the pumps. This kind of firmware sometimes does not like the compresibilty of the CO2. It is quite compressible compare to LC solvents.

5. Depending on the kind of autosampler, modification may need to be made. Loop filling styles seem to work better than "needle part of the loop" types. It's all about where any CO2 in the loop goes when the loop is decompressed. One option is to modify the AS to do what is known as "modifier stream injections", this can help.

6. Detector flow cells need to be modified to handle very high system pressures (4000 psi at least). Your average UV flow cell is usually rated at 1000 psi. Some flow cell gasket materials are compatible with CO2.

7. You would have to construct some kind of back pressure generating device after the detector to keep the CO2 liquid untill after it passes through the flow cell. A simple needle valve will often do the trick.


I didnt hit all the issues, it can be done if you are a bit of a mechanic. So, to simply answer your question, no you cant drag an old LC out of storage and hook the tank of CO2 up to it and run SFC.

Thank you both for the answers.
I should say what AA wrote is more than I can read in a textbook. Thanks again.
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry