Mass Spec Ion Source Parts
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
4 posts
Page 1 of 1
Is there ever a time to replace ion source parts other than a broken or worn connector?
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
Some ion sources have heaters; and heaters can fail. Filaments are used up. Parts can be broken or lost. But other than that, I don't recall seeing parts replaced.
Having said that - I work with GC systems and there may be some part in LCMS that I've not seen.
Having said that - I work with GC systems and there may be some part in LCMS that I've not seen.
-
- Posts: 196
- Joined: Mon Feb 22, 2010 11:15 am
In LC-MS you sometimes have to replace the probe capillary (if working in ESI mode) which can get clogged. If there is a cone/fragmentor at the entrance of the MS, you may want to replace it in case it is too dirty or broken. Otherwise there is little to do but frequently cleaning the source, as long as heating and voltages are ok.
-
- Posts: 92
- Joined: Sat Mar 01, 2008 12:31 pm
Thank you.
Other than deeply scratched parts I don't change out many ion source parts. I have fine grit paper for polishing the parts if they scratch too much and can get them to a mirror finish. Mostly the repeller and drawout plate.
The electrical connector pin will wear but I know that can be replaced if so desired.
I figure the tune algorithm will take care of any changes in the shape of the parts.
Other than deeply scratched parts I don't change out many ion source parts. I have fine grit paper for polishing the parts if they scratch too much and can get them to a mirror finish. Mostly the repeller and drawout plate.
The electrical connector pin will wear but I know that can be replaced if so desired.
I figure the tune algorithm will take care of any changes in the shape of the parts.
4 posts
Page 1 of 1
Who is online
In total there are 8 users online :: 1 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider] and 7 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider] and 7 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science