Posted: Thu Jul 01, 2010 7:14 pm
by Lukasz
My problem has begun 30 days ago. When I turn on my HPLC I observe to low pressure and large pressure fluctuations. My pressure should be 9 MPa, but when I turn on HPLC, pressure is only 4 or 5 MPa and it fluctuate about 20-30%. Of course I start to purge the pump, but it takes me 1 or sometimes 2 hours. When venting is finally finished pressure is very stable, fluctuate less than 2% and pressure is 9 MPa. Earlier I didn’t have such problems, purging was necessary only after, for example, 10 day long brake in apparatus usage
I have on line solvent degasser.
Please help me resolve my problem.

Posted: Thu Jul 01, 2010 9:06 pm
by danko
The degasser needs mending.

Better yet: Toss it away and degass your mobile phase prior to use.

Best Reards

Posted: Fri Jul 02, 2010 5:08 am
by Lukasz
Thank you for your answer. But I have doubt.
If degasser is damaged, why after long pump purging pressure back to normal?

Posted: Fri Jul 02, 2010 6:46 am
by danko
Well, then we’ll have to remove your doubts :wink:
In the beginning the air bubbles released in the pump heads and elsewhere in the system are very tiny. Over time they “find friendsâ€