- Posts: 15
- Joined: Mon Jun 15, 2009 10:39 pm
I am developing a drug confirmation method (in urine) on an Agilent 6400 LC/MS/MS. The method is for benzodiazepines and opioids and there are hydrolysis, extraction, and filtration procedures. I have noticed recently a hump (~1000 counts) near the end of the run (9 min), which I think is a result of a few benzos sticking to the column (oxazepam, temazepam, lorazepam, as I can tell from the MRM transitions). It is interfering with my recovery study (the recovery for my six replicates gradually increases for the compounds mentioned above). I have increased the time at the end of the run for which I hold at 100% acetonitrile. I tried rinsing with 100% organic three days ago, but this didn't help. Over the weekend I tried washing with 20 cvs (column volumes) ACN, 5 cvs IPA, 20 CVs hexane, 5 cvs IPA, and back to 20 cvs ACN, but to no avail. I backflushed the column (Zorbax C-18 2.1 x 100mm 1.8 um) this morning at 0.1 mL/min with ACN. When this didn't work I backflushed with IPA, and this still hasn't gotten rid of the hump. I run at 60 degrees C. Does anyone know of a wash program that might remove these benzodiazepines from this column? I will probably have to start using a new column, but I thought I would check to see if anyone had any suggestions first. Any help you can provide would be greatly appreciated.