FID Flame out Problem

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hi
I am facing a Flame out problem with my FID of Varian-450C at normal conditions but if we increase the Hydrogen flow it will work fine

So you should check the flowrate of H2 with a bubble flowmeter to see if it's correct.

Typical Detector FID settings are: 10% Hydrogen; 90% Air no?

If the FID is requiring more Hydrogen than normal,then problem is likely indicating that your glow-plug is about to burn out.
Jumpshooter

If the measured flows are correct the next step is to determine whether the flame is really out, or if the electyronics are indicating flame out when the flame is actually burning. Hold a cold shiny metal or glass surface over the detector, if you see condensation the flame is burning.

I have had two Varians that gave spurious flame out warnings when the instrument chassis picked up stray voltage from home made hardware. Do you have anything like a headspacer etc connected to the GC ?

Peter
Peter Apps
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 0 registered, 0 hidden and 2 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 2 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry