How often do you calibrate your HPLC? annually? quarterly?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
A bit of terminology here: "calibration" is done at least daily for a given analysis. It consists of establishing the proportionality between response and analyte amount.
In a regulated environment (e.g., pharmaceuticals), any valid method will also have "system suitability" tests which must be met before the method can be run on real samples.
System suitability overlaps a bit with "performance qualification" (which I think is what you are actually referring to). This involves running a defined series of tests on the chromatograph as a system, and often on the individual components. The frequency should be a matter of your SOPs (Standard Operating Procedures), and is typically at least once per year, and certainly after any maintenance or service operations have been performed on the system.
In a regulated environment (e.g., pharmaceuticals), any valid method will also have "system suitability" tests which must be met before the method can be run on real samples.
System suitability overlaps a bit with "performance qualification" (which I think is what you are actually referring to). This involves running a defined series of tests on the chromatograph as a system, and often on the individual components. The frequency should be a matter of your SOPs (Standard Operating Procedures), and is typically at least once per year, and certainly after any maintenance or service operations have been performed on the system.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
For us, PM/PV once a year by contract engineer, and daily system suitability injections and calculations for our cGMP and GLP work.
-
- Posts: 120
- Joined: Fri Jan 26, 2007 5:26 pm
Our HPLCs are requalified every six months by Agilent.
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
Annual PM and performance verification (calibration check), system suitability (at least standard RSD, if not resolution and tailing as well) on every run.
Thanks,
DR

DR

-
- Posts: 129
- Joined: Thu Oct 04, 2007 7:16 pm
PM/PV once a year by vendor, and system suitability injections with every run.
-
- Posts: 134
- Joined: Sat Dec 16, 2006 7:35 pm
PM/PV every 6 months minimum performed by in-house staff, system suit every run. Vendor only called if major breakdowns occurs.
7 posts
Page 1 of 1
Who is online
In total there are 35 users online :: 1 registered, 0 hidden and 34 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 34 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 34 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science