why bubble appear in Teflon tubing

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

12 posts Page 1 of 1
why bubble would appear in Teflon tubing between solvent bottle and pump head? what cause this? my flow rate is 18ml/min. is this flow rate too high for frit?
Thanks a lot if you can help.

The short answer: teflon is permeable to air. That's one of the advantages of the in-line degassers: they are located immediately adjacent to the pump inlet, so there is little opportunity for air to diffuse back in.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
why bubble would appear in Teflon tubing between solvent bottle and pump head? what cause this? my flow rate is 18ml/min. is this flow rate too high for frit?
Thanks a lot if you can help.
18 mL/Minute, what kind of chromatography are you doing :shock:
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
prep scale.
why bubble would appear in Teflon tubing between solvent bottle and pump head? what cause this? my flow rate is 18ml/min. is this flow rate too high for frit?
Thanks a lot if you can help.
18 mL/Minute, what kind of chromatography are you doing :shock:

thanks. I also used inline degasser. but it is for analytical usage.
and I do not think there is prep scale degasser as I saw in other thread.
The short answer: teflon is permeable to air. That's one of the advantages of the in-line degassers: they are located immediately adjacent to the pump inlet, so there is little opportunity for air to diffuse back in.

Just remove the frit and see whether you still get bubbles.

If the bubbles are really causing problems, use a bubble trap located just upstream of the pump inlet.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

thanks .
Just remove the frit and see whether you still get bubbles.

thanks.
If the bubbles are really causing problems, use a bubble trap located just upstream of the pump inlet.

Jiang295,

There are two reasons for seeing the bubbles in the Teflon tubings.

1. These are the only transparent components in the flow path allowing visibility of what’s in them (apart from the reservoirs if transparent and I bet you can see air bubbles in them too).

2. Teflon is quite hydrophobic and air “feelsâ€
Learn Innovate and Share

Dancho Dikov

hi danko,

I did use vacuum plus ultrasonic to degass our sample.

[quote="danko"]Jiang295,

There are two reasons for seeing the bubbles in the Teflon tubings.

1. These are the only transparent components in the flow path allowing visibility of what’s in them (apart from the reservoirs if transparent and I bet you can see air bubbles in them too).

2. Teflon is quite hydrophobic and air “feelsâ€

hi danko,

I did use vacuum plus ultrasonic to degass our sample.
And now you might like to degass the mobile phase as well.

Best Regards
Learn Innovate and Share

Dancho Dikov
12 posts Page 1 of 1

Who is online

In total there are 14 users online :: 3 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], bubblegum, Semrush [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry