18 mL/Minute, what kind of chromatography are you doingwhy bubble would appear in Teflon tubing between solvent bottle and pump head? what cause this? my flow rate is 18ml/min. is this flow rate too high for frit?
Thanks a lot if you can help.
18 mL/Minute, what kind of chromatography are you doingwhy bubble would appear in Teflon tubing between solvent bottle and pump head? what cause this? my flow rate is 18ml/min. is this flow rate too high for frit?
Thanks a lot if you can help.
The short answer: teflon is permeable to air. That's one of the advantages of the in-line degassers: they are located immediately adjacent to the pump inlet, so there is little opportunity for air to diffuse back in.
Just remove the frit and see whether you still get bubbles.
If the bubbles are really causing problems, use a bubble trap located just upstream of the pump inlet.
And now you might like to degass the mobile phase as well.hi danko,
I did use vacuum plus ultrasonic to degass our sample.