flow rate(He) of GC/MS for different inner diameter?

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hi, everyone!
I use aglient 7890-5975C GC/MS. I want to know how to set flow rate(He) of GC/MS for different inner diameter capillary column?
Thanks!

The flowrate in GC/MS is determined by MSD which is appr. 1.0ml/min (can be 1.5) for best performance. So you can not use megabore columns (eg. 0.53mm). Even 0.32 mm column is not good (and not possible for short columns).

Ideal linear velocity in a column is about 23 cm/sec. With a 0.32 column, you can do this, but the column needs to be long enough that the pressure drop will allow the inlet to be operated at above atmospheric pressure. If running your column at optimum velocity delivers more gas to the mass spec than the pump can handle, you need a narrower column. Pumping capacity of a mass spec is determined by the pump in use on the mass spec. Check the specifications.
3 posts Page 1 of 1

Who is online

In total there are 28 users online :: 0 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry