Does it happen with another batch of acetone?
Does it happen with any other analyte?
If you change the voltage of the skimmer (Agilent calls it fragmentor, Waters cone voltage), do you see any effect?
I did use acetone and besides schiff bases with my analytes I couldn't see anything special.
I can't say whether we evaluated different batches of acetone. However, this same adduction was seen for paclitaxel, paclitaxel-d5 and 6-hydroxypaclitaxel. The Sciex uses declustering potential for up-front CID and yes, this did have an effect on the abundance of the adduct. However, at the optimal sensitivity for the [M+H]+ the adduct was approximately the same abundance so we never eliminated it in this manner (though we didn't necessarily need to).
The Schiff base thought is interesting. So were you seeing M+H+40? Hopefully it wasn't M+40 or I'm left asking the same question about how an ion is being formed.