Mobile phase consists of 15 % MeOH, 0.1 M citrate buffer pH 4.5, 0.6 mM octanesulphonic acid and 1.2 mM EDTA. Flow rate is 1 ml/min We usually have mobile phase constantly circulating in the system (at low 0.2 ml/min flow rate when not in use) so as to cut down the time needed for equilibrating.
The latest problem with this instrument is the weird dopamine peak that comes and goes. I injected dopamine standards and managed to see a peak that seemed to be dopamine (retention time around 10 min). Shape of the peak was good. However, when injected again after some time (10-20 min) the peak had vanished. I suspected that dopamine was eluting this time right after the solvent front but could not be sure. Third injection produced no peaks at all other than the solvent front. The sample was kept in ice at all times so it's very unlikely that it would have degraded between the analyses. I was using a very concentrated standard so if there would have been a peak it would have been clearly visible. I redid this experiment on another day and again experienced an oddly "bouncing" peak. If I just let the instrument run and chartwriter register without injections or if I do a blank injection there are no peaks.
Solvent front peaks always appear 2.4 min after the injection, so the injections seem to be succesful. I also tested the pump and it seems to be working ok. Where is the problem? Injector still failing somehow? Column issue? Detector hiccups?
I really appreciate if somebody could help me with suggestions what could be checked or improved. Thank you!