broad solvent front in GC
Discussions about GC and other "gas phase" separation techniques.
4 posts
Page 1 of 1
I have always run a toluene blank with a 500 sec. acquisition delay on the detector. After this delay, I would naturally have no toluene show up as it would have all been gone by then. I ran a sample which contained 200 pg of analyt - and i injected 5 ul. Now since then, I'm getting my solvent peak tailing off for the entire run ! Even after about 500 seconds delay the peak abundance is up to 120,000 units. any suggestions?
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
and / or
column trimming or replacement.
Rodney George
column trimming or replacement.
Rodney George
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
I would also look at the split valve. If not operating correctly and you are staying splitless you sensitivity would not improve but your solvent peak will go on for the full run.
Best regards.
Best regards.
4 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science