A time of flight spectrometer pulses ions into the flight tube at a phenominal rate - like 5000 times per second, but the signals are summed and sums of several acquisitions are delivered to the use as a spectrum. I've used the LECO system for a number of years and it acquires 5000 spectra per second - but give acquisition rates of 500 or lower spectra per second to the user. So, you have a sum of at least 10 "spectra" to give a spectrum that you see. I have never used that rate for acquisition - but watch out. An acquisition rate of 2 Hz will only give you 20 data points across a 10 second wide peak (at base). If you use any fast chromatographic techniques, this will limit how fast you can run.
With all of the things that get done by the software and hardware, ask each vendor to show you the results of running a typical sample. It is best to have a sample that you know and understand. And, ask to talk with a a customer or two who are failry close by. See how well the organization supports the instruments when they get into the customers' labs.
With all of the things that get done by the software and hardware, ask each vendor to show you the results of running a typical sample. It is best to have a sample that you know and understand. And, ask to talk with a a customer or two who are failry close by. See how well the organization supports the instruments when they get into the customers' labs.
-
- Posts: 198
- Joined: Tue Mar 27, 2007 6:54 pm
I am looking at purchasing either an Almsco Bench TOF-dx or a Waters GCT Premier.
Comparing the two has been ok apart from one aspect, the Almsco quotes "Data range: 5000 spectra per data point (2 Hz)" in some application notes whereas the Waters quotes "a spectral sampling rate of up to 20 spectra per second".
My question is which is better? What exactly are Almsco quoting here?
Has anyone any experience with either system?
Thanks in advance.
Gcmol,
I don't have neither of them but in these days I'm looking in that direction too.
These are two different systems. The Almsco is a low resolution (probably similar to a LECO) the Waters is high resolution. What are your plans for a GC-TOF? Characterization, GCxGC?
Regards,
bhuvfe
-
- Posts: 12
- Joined: Tue May 25, 2010 11:41 pm
solved
-
- Posts: 198
- Joined: Tue Mar 27, 2007 6:54 pm
Then, as Don suggested, a demo on both is the best way to go... (even because one supplier uses OFN the othe hexachlorobenzene in their specs).
Have you considered a Almsco + Massworks post-processing software?
Have you considered a Almsco + Massworks post-processing software?
-
- Posts: 8
- Joined: Tue Jun 08, 2010 1:01 pm
Hi all,
I came across this paper in CTMS where they use the Almsco Bench TOF-dx for GCxGC analysis. In it they state a data rate of 50Hz and the chromatograms look pretty good.
http://chromatographyonline.findanalyti ... ail/667522
I guess different chromatographic separations require different data rates.
Hope this helps!
Dan
I came across this paper in CTMS where they use the Almsco Bench TOF-dx for GCxGC analysis. In it they state a data rate of 50Hz and the chromatograms look pretty good.
http://chromatographyonline.findanalyti ... ail/667522
I guess different chromatographic separations require different data rates.
Hope this helps!
Dan
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
Note that the article cited is written by a person working for the vendor. Not that this is bad, but don't expect much discussion on limitations. The claim that tradidional TOF instruments do not produce spectra classic spectra that can be searched in libraries is bogus. I've been doing it for years on a GC-TOF. With that claim I'd be careful of the "facts" given. (I have to assume that the author of the article has confused spectra from differnt types on ion sources, such as those seen in LC/MS into the mix - but GC-TOF instruments typically use EI, just like quads, sectors, etc.)
Do your homework. Talk to prospective vendors. Let them tell you the advantages over other instruments, but check to see if those advantages are real.
Do your homework. Talk to prospective vendors. Let them tell you the advantages over other instruments, but check to see if those advantages are real.
-
- Posts: 20
- Joined: Fri Mar 31, 2006 11:47 am
There's no need to erase the original question if it has been solved. I've found the answers to many questions just by searching to see if it has been asked before (and it usually has). It's a great forum and source of information.solved
cheers
Pete
8 posts
Page 1 of 1
Who is online
In total there are 61 users online :: 0 registered, 0 hidden and 61 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 61 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 61 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science