Hi all...
I have questions about column efficiency.
When I use acetonitrile and methanol as mobile phase, the efficiency is about 2000,
but when I introduce water to the mobile phase, the efficiency fell to 500, and my analyte peak became short and wide..
How to increase the efficiency so I can get narrow and tall peak? I have to use water in mobile phase to separate my analyte from the matrix. The concentration of water I used is about 15%.

Thanks for your help.. :)
It would be helpful to know what column you are using, RP, HILIC or which one.
Gerhard Kratz, Kratz_Gerhard@web.de
Also the column dimensions, packing particle size, flow rate, injection volume, and diluent. All of those have an effect.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Are you using a sysyem capable of running gradients?
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
@ Gerhard & Tom:
I used RP column, Zorbax Eclipse XDB C18 150mm x 4.6, 5 micron.
I use 1.2 ml/minute flow rate, 20 micron injection volume and acetonitrile as solvent.
I use metahnol :water = 85% : 15% as mobile phase, Uv detection at 210nm, and 40°C column temperature.
I have changed the Zorbax column with Hypersil ODS, 125mm x 4.0, 5 micron, but the efficiency still low.

@ JGK:
Yes, it is possible to run gradient analysis
First of all, even 2000 plates is low for this column. A 150 x 4.6 mm / 5 micron column should be able to generate 5-10,000 plates for small molecules. If nothing else, you should check your system for exessive extra-column volume (too large a tubing diameter; excessively long tubing; poorly assembled fittings).

Second, you did not tell us the approximate retention time of your analyte. If it is weakly retained (on that column at 1.2 mL/min, the dead time is around 1.2 minutes, give or take 15%) then plate count measurements are unreliable. To get reasonable results, you should measure on a peak whose k' is > 1 (i.e., > 2x the dead time).

Third, peak shape problems (including excessive broadening) are reasonably common when the diluent is stronger than the mobile phase (in your case, 100% ACN as the diluent vs. 85% MeOH in the mobile phase. For best results, it's a good idea for the diluent to match the mobile phase.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
6 posts Page 1 of 1

Who is online

In total there are 24 users online :: 1 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot] and 23 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry