Is in-line degasser recommended for semi-prep pumps?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
Hi all,

I was told that an in-line degasser is not essential in semi-prep HPLC. Is it because semi-prep is often used in isocratic mode and with organic solvents?

If I plan to do semi-prep reverse-phase gradients, is an in-line degasser as important as for analytical HPLC?

Thanks for your answers,

Best regards

That depends on what kind of pump you have.

The "degassing" problem stems from the fact that the solubility of air in mixtures of polar organic solvents + water is less that the average of the solubilities in the starting solvents. Therefore, when you mix methanol (for example) and water, the mixture is supersaturated with air, and the excess air "precipitates" as bubbles. If this happens upstream of the pump (as it would in a "low-pressure-mixing" system), then the bubbles can interfere with pump operation. If it happens downstream of the pump (as it would in a "high-pressure-mixing" system), then the bubbles can show up as spikes on the baseline.

So, if you are using a high-pressure mixing system, and you don't mind the spikes, then you can get by without degassing.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Thanks Tom, I am planning to buy a high pressure mixing binary pump. Indeed I could do without degassing since in a first step I will not have a detector other than bioactivity :D

but does it affect separation for high pressure mixing?
That depends on what kind of pump you have.

The "degassing" problem stems from the fact that the solubility of air in mixtures of polar organic solvents + water is less that the average of the solubilities in the starting solvents. Therefore, when you mix methanol (for example) and water, the mixture is supersaturated with air, and the excess air "precipitates" as bubbles. If this happens upstream of the pump (as it would in a "low-pressure-mixing" system), then the bubbles can interfere with pump operation. If it happens downstream of the pump (as it would in a "high-pressure-mixing" system), then the bubbles can show up as spikes on the baseline.

So, if you are using a high-pressure mixing system, and you don't mind the spikes, then you can get by without degassing.

but does it affect separation for high pressure mixing?
Probably not. The bubbles generally will only form once the mobile phase approaches atmospheric pressure (i.e., at the outlet of the column).

I can imagine an exception if you have analytes that are easily oxidized, in which case the dissolved oxygen might have an effect.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

thanks.
[
quote="tom jupille"]
but does it affect separation for high pressure mixing?
Probably not. The bubbles generally will only form once the mobile phase approaches atmospheric pressure (i.e., at the outlet of the column).

I can imagine an exception if you have analytes that are easily oxidized, in which case the dissolved oxygen might have an effect.[/quote]
6 posts Page 1 of 1

Who is online

In total there are 19 users online :: 2 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry