I wondered if anyone can provide an answer the issue I am currently having, I am running a Chiral method, which requires derivatisation.
My sample is heat sensitive so the autosampler is 10C.
My mobile phase is phosphate buffer pH3.0 / methanol and the sample diluent aqueous pH3.0.
Since I am doing method development I am worried about the instrument, and have yet to complete SST to assure the system is injecting reproducibly.
What areas would please suggest I look at, needle and needle wash?
Could it be a leak from the needle wash as opposed to sample leak from the needle?
FYI Instruments is a Waters 2695.
Thanks for any tips!