Loss of Bromoform during VOA analysis
Posted: Mon Jun 07, 2010 2:46 pm
by markeith6172
I am currently using an Aechon autosampler with a OI 4660 P&T connected to an agilent 6880GC and 5273MS. My issue is that we have recently lost sensitivity for bromoform. We have changed the sparger, transfer line from the auto sampler to the P&T and to the GC. We have also change the trap (#10) but bromoform is only recovering about 30%. Any suggestions would be appreciated. Thanks
Posted: Mon Jun 07, 2010 5:16 pm
by Steve Reimer
Check your purge flow. Low Bromoform is a sign of incomplete purge.
Also, if you haven't done so yet, run the part b leak check on your 4660.
Posted: Mon Jun 07, 2010 5:17 pm
by Steve Reimer
Re: Loss of Bromoform during VOA analysis
Posted: Mon Jun 07, 2010 7:02 pm
by Caltech
[quote="markeith6172"]I am currently using an Aechon autosampler with a OI 4660 P&T connected to an agilent 6880GC and 5273MS. My issue is that we have recently lost sensitivity for bromoform. We have changed the sparger, transfer line from the auto sampler to the P&T and to the GC. We have also change the trap (#10) but bromoform is only recovering about 30%. Any suggestions would be appreciated. Thanks[/quote]
Do you ever heard Metal quad analyzer and high performance ion source design by caltechscientific.com .... specially could solve the problem for you... or call help line at 650-799-4230
Henry
Posted: Wed Jun 09, 2010 8:14 pm
by Ron
You may be losing bromoform in internal tubing inside the purge and trap or in the water managment fitting. I would suggest taking out the the water management fitting and sonicating in methanol. Dry in a oven for about an hour, reinstall, and see if the bromoform is back. If this doesn't help consider carefully removing the internal lines going from the sparge mount to the 6-port valve, 6-port to trap, and flush them with methanol, oven dry, and reinstall.
Are you using a 5 mL or 25 mL sparge volume?