Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
4 posts
Page 1 of 1
I need to buy an used LC/MS/MS for quantification of several biomarkers. Is API 3000 more sensitive? According to its datasheet of API 3000, 0.8 pg of a compound can have a peak with S/N of 150, while for Quantum Classic, 10 pg of reserpinep produce a S/N ratio of 150.
-
- Posts: 1889
- Joined: Fri Aug 08, 2008 11:54 am
Beware of comparing S/N values from different manufacturers without checking that they are calculated in exactly the same way. Even within one software package (Xcalibur from Thermo), different integrators can produce S/N values for the same peak that differ by orders of magnitude.
Companies also differ in their optimism when they write specification sheets!
If you are buying an LC-MS, the manufacturers should be prepared to run some test samples for you. Make up an appropriate standard and matrix, and ask them to give you limit of quantification data by some reasonably standard method. Direct comparison of real data from your samples will be more reliable than a specification sheet.
Companies also differ in their optimism when they write specification sheets!
If you are buying an LC-MS, the manufacturers should be prepared to run some test samples for you. Make up an appropriate standard and matrix, and ask them to give you limit of quantification data by some reasonably standard method. Direct comparison of real data from your samples will be more reliable than a specification sheet.
-
- Posts: 8
- Joined: Fri Jan 08, 2010 7:27 pm
Thanks a lot.Beware of comparing S/N values from different manufacturers without checking that they are calculated in exactly the same way. Even within one software package (Xcalibur from Thermo), different integrators can produce S/N values for the same peak that differ by orders of magnitude.
Companies also differ in their optimism when they write specification sheets!
If you are buying an LC-MS, the manufacturers should be prepared to run some test samples for you. Make up an appropriate standard and matrix, and ask them to give you limit of quantification data by some reasonably standard method. Direct comparison of real data from your samples will be more reliable than a specification sheet.
-
- Posts: 1889
- Joined: Fri Aug 08, 2008 11:54 am
sorry, didn't spot of course that this is a used instrument, but it may (should?) still be possible to get a very simple demo to check S/N on low concentration injections of a compound of your choice.
4 posts
Page 1 of 1
Who is online
In total there are 3 users online :: 0 registered, 0 hidden and 3 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 3 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 3 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science