Foreign late SEC peaks
Posted: Thu Jun 03, 2010 9:13 pm
Hello Everyone,
We are running a Waters 2695 solvent/sample management system with a styragel column bank and toluene as the mobile phase. We have identical columns with THF mobile phase as well. Recently we have noticed carryover peaks from sample to sample. I ran a longer than usual sample (our injections usually run for an hour) to see when this late peaks (magenta) came out and noticed that they come out much later than our standard solvent peaks (red). If we run the same samples but replacing the mobile phase to THF and switching the columns we do not see these very late peaks. Can anyone suggest possible reasons for these?
Some questions/thoughts:
1. Can this be due to defective columns (please NO!)
2. It cannot be due to impure solvent as we have gone through different HPLC grade solvents and we see the same thing.
An easy solution is to wait until they come out for the next injection but we would rather not waste that amount of solvent.
Thanks for your help!