Dionex Carbohydrate Troubleshooting
Posted: Thu Jun 03, 2010 4:13 pm
Hello Everyone,
I'm an annoyed novice when it comes to Dionex. Unfortunately, I have to use the equipment to analyze some carbohydrates. So here's my current problem:
The chromatography for my sugar standards are bad. You can't see the peaks, the ones you do see aren't repeatable, and the baseline is really high.
I ran this analysis back in December and everything worked out with the standards. Before I shut down the system (until now), I purged the column with 150mM NaOH and then capped it. I then flushed 18M-Ohm water through the pump and the electrodes. I then stored the reference electrode in its KI solution.
Before I ran the standards yesterday, here's what I did:
1) Collected fresh 18M-Ohm water. Purged with He for 30 min. (Eluent A)
2) Made fresh 300mM NaOH. Filtered with a 0.2um nylon and purged with He for 30 min. (Eluent C)
3) Both Eluents are stored under blanketed Helium.
4) Flushed the system with water (no column in) for 30 min.
5) Calibrated my reference electrode- went well.
6) Polished my working electrode (Gold, non-disposable).
7) Uncapped column.
Changed column filters in guard (both) and analytical (front). The guard was a bit orange brown in the beginning, but the end was the nice cream color. The analytical was nicely cream.
9) Connected freshly-trimmed black tubing from guard to analytical.
10) Hooked the columns to the system (not the cell).
11) Flushed with Eluent A (water) to remove any junk (30 min).
12) Attached columns to cell.
13) Using Manual Acquisition to collect the baseline, ran the system at starting conditions (100% Eluent A- water) for about an hour.
***This is where the problems started. It looks like for about 20 minutes I had a nice baseline, but from 20 to 60 minutes, the baseline just kept creeping up.***
Concerned with this, thinking that I futzed with the column by using water, I ran the standards overnight hoping to see an improvement over the course of the run.
Unfortunately, there was no such luck. The chromatograms look just as bad as they did in the beginning. I'm trying to figure out what I did wrong, or if it's the column. If it is the column, do I need to regenerate or purchase a new one? The data is due on Monday and my time is running out.
Any help you guys could give would be great. From my previous calls with Dionex, it's hard to find someone who can actually help you. Thanks again!
I'm an annoyed novice when it comes to Dionex. Unfortunately, I have to use the equipment to analyze some carbohydrates. So here's my current problem:
The chromatography for my sugar standards are bad. You can't see the peaks, the ones you do see aren't repeatable, and the baseline is really high.
I ran this analysis back in December and everything worked out with the standards. Before I shut down the system (until now), I purged the column with 150mM NaOH and then capped it. I then flushed 18M-Ohm water through the pump and the electrodes. I then stored the reference electrode in its KI solution.
Before I ran the standards yesterday, here's what I did:
1) Collected fresh 18M-Ohm water. Purged with He for 30 min. (Eluent A)
2) Made fresh 300mM NaOH. Filtered with a 0.2um nylon and purged with He for 30 min. (Eluent C)
3) Both Eluents are stored under blanketed Helium.
4) Flushed the system with water (no column in) for 30 min.
5) Calibrated my reference electrode- went well.
6) Polished my working electrode (Gold, non-disposable).
7) Uncapped column.
9) Connected freshly-trimmed black tubing from guard to analytical.
10) Hooked the columns to the system (not the cell).
11) Flushed with Eluent A (water) to remove any junk (30 min).
12) Attached columns to cell.
13) Using Manual Acquisition to collect the baseline, ran the system at starting conditions (100% Eluent A- water) for about an hour.
***This is where the problems started. It looks like for about 20 minutes I had a nice baseline, but from 20 to 60 minutes, the baseline just kept creeping up.***
Concerned with this, thinking that I futzed with the column by using water, I ran the standards overnight hoping to see an improvement over the course of the run.
Unfortunately, there was no such luck. The chromatograms look just as bad as they did in the beginning. I'm trying to figure out what I did wrong, or if it's the column. If it is the column, do I need to regenerate or purchase a new one? The data is due on Monday and my time is running out.
Any help you guys could give would be great. From my previous calls with Dionex, it's hard to find someone who can actually help you. Thanks again!