Pressure fluctuations and air bubbles

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
We are experiencing some problems with our Agilent 1200 Quaternary pump/1200 series vacuum degasser. We experienced contamination problems with the degasser and it was taken off line. Helium was used to degas the solvents [acetonitrile (pump B)/water (pump A) used in a gradient]. The pressure began fluctuating and air bubbles were observed going from the MCGV valve to the pump. The bubbles were also coming out of the detector and spikes were observed on the baseline. When the degasser was put back on line, the problem went away. Any ideas as to what may be causing these pressure fluctuations and bubbles when the degasser is replaced with helium sparging?

thanks

It sounds like sparging was done done sufficiently, or that air was able to get into the lines when you reconnected after removing the degasser...
2 posts Page 1 of 1

Who is online

In total there are 42 users online :: 2 registered, 0 hidden and 40 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 40 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry