My hplc condition is 20mmol KH2PO4 (PH=3.2):acetonitrile=96.3:3.7 ,which is operated by two pumps.when i add a new batch of mobile phase ,the retention of my object would change a little and during the running time ,it changed as well.My column is C18,250*4.6,5,intex.because the condition is used to detect the serum samples by hplc-uv,a little change in retention will bring me into troubles.In my opinion,the chromatograhy method may be not rough enough,but my question is that whether the column stable enough in such conditon.thanks a lot!
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
You know, it would have made more sense to simply continue your previous thread on this topic:
viewtopic.php?t=13033&highlight=
That said, most C18 columns should be reasonably stable at pH 3.2. Depending on your sample prep, garbage from the serum may be accumulating and changing the column chemistry. If that's the case, a better sample cleanup is the best remedy. The use of a guard cartridge can also help, but you may have to change it frequently.
viewtopic.php?t=13033&highlight=
That said, most C18 columns should be reasonably stable at pH 3.2. Depending on your sample prep, garbage from the serum may be accumulating and changing the column chemistry. If that's the case, a better sample cleanup is the best remedy. The use of a guard cartridge can also help, but you may have to change it frequently.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 13
- Joined: Tue May 25, 2010 9:02 am
I really appreciate your reply!thanks very much!
-
- Posts: 13
- Joined: Tue May 25, 2010 9:02 am
I am also confused that my condition which contains only 3.7% acetonitrile,is it one of the cause of unstable method?
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
If it only takes 3.7 % ACN to elute your analyte from a C18 column, maybe it’s not the right technique for this application.
The above could also (to some degree) explain the retention instability you’re experiencing.
Best Regards
The above could also (to some degree) explain the retention instability you’re experiencing.
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 13
- Joined: Tue May 25, 2010 9:02 am
thanks a lot
6 posts
Page 1 of 1
Who is online
In total there are 14 users online :: 0 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 14 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 14 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science